Brian
GuestThe meter shows 0.00113 to 0.00095 A with no leads connected. How can this be used to measure CPU deep sleep currents which are in the uA? Is calibration required?
Thanks,
Brianadmin
KeymasterHi Brian –
The meters are calibrated at the factory, but it sounds like you got a bad calibration. Typical offset is in the tens of microamps.
We’re happy to replace your meter, very sorry for the trouble. Please email at hello@moosh.im to arrange it.
Also there’s a “zero” button in the app that will zero out the DC offset, but it’s only a temporary offset. Your meter should not be reading 1mA off out of the box.
~James